Large Molecule View: European Bioanalysis Forum Feedback Joanne Goodman, on behalf of the EBF http://www.e-b-f..eu

Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

  • Upload
    others

  • View
    3

  • Download
    0

Embed Size (px)

Citation preview

Page 1: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Large Molecule View:European Bioanalysis Forum Feedback

Joanne Goodman, on behalf of the EBF

http://www.e-b-f..eu

Page 2: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Disclaimer

This presentation was prepared on behalf of EBF incorporating to the best of our ability the outcome of internal EBF discussions and surveys, the discussions from the EBF Barcelona Industry Focus Workshop (sister meeting) discussions.

The opinions expressed in this presentation do not necessarily reflect the view of any individual expert, any individual EBF member company, AstraZeneca or that of the ICH M10 Expert Working Group (EWG).

2

Page 3: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

To go forwards sometimes you need to step back ….

3

Page 4: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

The first important step …..

Ø 1990 - Analytical Methods Validation: Bioavailability, Bioequivalence and Pharmacokinetic Studies meeting of industry and regulators – Became known as the first Crystal City meeting– Bioanalytical method validation and performance mainly for HPLC

and GC – Summarised by Shah (1992)– First set of expectations for PK methods adopted by regulators– Basis of guidance today– However, we are now suffering from “Copy, Paste, Add”

4

Page 5: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

BMV guidance has evolved over time

5

1992 – HC BA and BE studies1997 – FDA 21 CFR Part 11, Japan conduct of nonclinical studies1998 – OECD for GLP2000 – Crystal City II2001 – FDA bioanalytical method validation guidance2003 – FDA BA and BE studies, ANVISA bioanalytical method validation guidance2005 – ANVISA regulation revision, India BA and BE guideline, ICH Q2(R1) validation of analytical techniques, CFDA BA and BE guideline2006 – Crystal City III 2007 – Crystal City IV2009 – WHO GCLP2010 – EMA BE guideline2011 – CFDA guidance on management of laboratory for clinical trial sample analysis, EMA guideline on bioanalytical method validation2012 – ANVISA RDC 27 resolution, HC comparative BA studies, EMA GCLP reflection paper2013 – MHLW bioanalytical method validation (chromatography), FDA draft guidance for bioanalytical method validation, Crystal City V2014 – MHLW bioanalytical method validation (LBA)2015 – EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation2018 – FDA BMV

Page 6: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Current guidance stemmed from chromatography and the small molecule perspective

Ø The early BMV guidance documents were small molecule/chromatographic focussedØ 2006 – CCIII that LBA becomes more prominent

Ø Ligand binding assays (LBA) were only covered at a superficial level– Some guidance also applies to other bioanalytical methods, such as

immunological and microbiological procedures…– “Chromatography creep”

6

Page 7: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

But remember …..

7

Chromatography Assays

¹Ligand Binding Assays

Page 8: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Other considerations for LBA

Ø More new modalities for biologicsØ Different technologies and techniquesØ But not everything sits nicely in technology buckets

– Large molecules being analysed by chromatography – “Hybrid” assays– The future state ……

8

Page 9: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Does the M10 draft guideline create alignment?

9

Page 10: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Does the M10 draft guideline create alignment?

10

ICH's mission is to achieve greater harmonisation worldwide to ensure that safe, effective, and high quality medicines are developed and registered in the most resource-efficient manner

(source: www.ich.org)

Page 11: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Do we all see the guidance in the same way?

11

Page 12: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

EBF Feedback from Barcelona: Ligand Binding Assays

12

Page 13: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

LBA topics in ICH M10

13

Page 14: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

To prevent overload ……

Ø Focus on the major topics for the feedback from EBF– Some positive changes and

clarifications in ICH M10– However, there are those that will

bring challenges in our everyday activities for reasons of:oResourceoCostoTime

14

Page 15: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Format of the slides

Ø Black text is the ICH M10 text

Ø Blue text with a bold strike through is a suggested deletion

Ø Red text is the addition or changed language

15

Page 16: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Separate sections for LBA

Ø PositiveØ Removes potential ambiguity when looking at a cross-referenced

chromatography sections

16

Page 17: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Reference Standard

Ø Add: If the reference standard batch used for bioanalysis is changed, bioanalytical evaluation should be carried out prior to use to ensure that the performance characteristics of the method are within the acceptance criteria and to ensure consistency of results between batches in case of change during bioanalysis of samples from a given nonclinical or clinical study

17

It is recommended that the manufacturing batch of the reference standard used for the preparation of calibration standards and QCs is derived from the same batch of drug substance as that used for dosing in the nonclinical and clinical studies whenever possible. If the reference standard batch used for bioanalysis is changed, bioanalytical evaluation should be carried out prior to use to ensure that the performance characteristics of the method are within the acceptance criteria.

Page 18: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Critical Reagents

Ø Add/Delete: Critical reagents bind the analyte and, upon interaction, lead to an instrument signal corresponding to the analyte concentration. Critical reagents, including binding reagents (e.g. binding proteins, aptamers, antibodies or conjugated antibodies), have direct impact on the results of the assay and therefore their quality must be assured. The critical reagents should be identified and defined in the assay method. Reliable procurement of critical reagents, whether manufactured in-house or purchased commercially, should be considered early in method development.

18

Critical reagents, including binding reagents (e.g., binding proteins, aptamers, antibodies or conjugated antibodies) and those containing enzymatic moieties, have direct impact on the results of the assay and, therefore, their quality should be assured. Critical reagents bind the analyte and, upon interaction, lead to an instrument signal corresponding to the analyte concentration. The critical reagents should be identified and defined in the assay method. Reliable procurement of critical reagents, whether manufactured in-house or purchased commercially, should be considered early in method development.

Page 19: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Critical Reagents

Ø Minor and major changes Ø Use of re-test dates rather than inflexible expiry dates Ø Positive changes

19

Page 20: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Singlicate vs. duplicate wells

Ø Positive change

20

When using LBA, study samples can be analysed using an assay format of 1 or more well(s) per sample. The assay format should be specified in the protocol, study plan or SOP. If method development and assay validation are performed using 1 or more well(s) per sample, then study sample analysis should also be performed using 1 or more well(s) per sample, respectively. If multiple wells per sample are used, the reportable sample concentration value should be determined either by calculating the mean of the responses from the replicate wells or by averaging the concentrations calculated from each response. Data evaluation should be performed on reportable concentration values.

Page 21: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Specificity

Ø Insert definition of what consists a ‘structurally related molecule’

Ø The accuracy of the target analyte at the LLOQ and at the ULOQ High QC should be investigated in the presence of related molecules at the maximal concentration(s) anticipated in study samples. The response of blank samples spiked with related molecules should be below the LLOQ. The accuracy of the target analyte in presence of related molecules should be within ±25% for the LLOQ and ±20% of the nominal High QC spike respectively.

21

Specificity is evaluated by spiking blank matrix samples with related molecules at the maximal concentration(s) of the structurally related molecule anticipated in study samples.

The accuracy of the target analyte at the LLOQ and at the ULOQ should be investigated in the presence of related molecules at the maximal concentration(s) anticipated in study samples. The response of blank samples spiked with related molecules should be below the LLOQ. The accuracy of the target analyte in presence of related molecules should be within ±25% of the nominal values.

Page 22: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Specificity (continued)

Ø .....It is essential to determine the minimum concentration of the related molecule where interference occurs If relevant, a minimum concentration of the related molecule causing interference should be determined at the maximal concentration of the target analyte

22

In the event of non-specificity, the impact on the method should be evaluated by spiking increasing concentrations of interfering molecules in blank matrix and measuring the accuracy of the target analyte at the LLOQ and ULOQ. It is essential to determine the minimum concentration of the related molecule where interference occurs. Appropriate mitigation during sample analysis should be employed, e.g., it may be necessary to adjust the LLOQ/ULOQ accordingly or consider a new method.

Page 23: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Selectivity – lipaemic and haemolysed

Ø Recommended that for cause activityØ Rephrase: Selectivity should be evaluated in lipaemic samples and haemolysed

samples if required. For lipaemic and haemolysed samples, tests can be evaluated once using a single source of matrix as part of the 10 individuals.

Ø Rephrase: In the case of relevant patient populations, when available, there should be at least five individual patients.

Ø Add text from Section 3.2.1 rather than a reference, including the suggested edits to the original draft

23

Selectivity should be evaluated in lipaemic samples and haemolysed samples (Refer to Section 3.2.1). For lipaemic and haemolysed samples, tests can be evaluated once using a single source of matrix. Selectivity should be assessed in samples from relevant patient populations. In the case of relevant patient populations there should be at least five individual patients.

Page 24: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Calibration Curve and Range

Ø Add: Calibration standards and QCs prepared in a matrix different from the study samples should be justified and appropriate experiments should be performed

Ø Clarify: LLOQ may not be the lowest standard and the ULOQ may not be the highest standard due to anchor points

24

The calibration curve demonstrates the relationship between the nominal analyte concentration and the response of the analytical platform to the analyte. Calibration standards, prepared by spiking matrix with a known quantity of analyte, span the calibration range and comprise the calibration curve. Calibration standards should be prepared in the same biological matrix as the study samples. The calibration range is defined by the LLOQ, which is the lowest calibration standard, and the ULOQ, which is the highest calibration standard. There should be one calibration curve for each analyte studied during method validation and for each analytical run.

Page 25: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Calibration Curve and Range (continued)

Ø Clarify “calibration curve parameters” regarding the blank sample

25

A calibration curve should be generated with at least 6 concentration levels of calibration standards, including LLOQ and ULOQ standards, plus a blank sample. The blank sample should not be included in the calculation of calibration curve parameters. Anchor point samples at concentrations below the LLOQ and above the ULOQ of the calibration curve may also be used to improve curve fitting. The relationship between response and concentration for a calibration curve is most often fitted by a 4- or 5-parameter logistic model if there are data points near the lower and upper asymptotes, although other models may be used with suitable justification.

Page 26: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Calibration Curve and Range (continued)

Ø Rephrase: A minimum of 6 independent runs should be evaluated over several 2 or more days considering the factors that may contribute to between-run variability.

Ø Rephrase: If freshly spiked calibration standards are not used, the frozen calibration standards can be used within their defined period of stability under the same conditions as freeze-thaw stability and a defined period of stability has been proven for the calibrator standards.

26

A minimum of 6 independent runs should be evaluated over several days considering the factors that may contribute to between-run variability.

The calibration curve should preferably be prepared using freshly spiked calibration standards. If freshly spiked calibration standards are not used, the frozen calibration standards can be used within their defined period of stability.

Page 27: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Preparation of Quality Control Samples

Ø Add: Calibration standards and QCs prepared in a matrix different from the study samples should be justified and appropriate experiments should be performed.

27

The QCs are intended to mimic study samples and should be prepared by spiking matrix with a known quantity of analyte, stored under the conditions anticipated for study samples and analysed to assess the validity of the analytical method.

Page 28: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Preparation of Quality Control Samples (continued)

Ø Clarify: “preparation from a single stock”Ø Add: definition of stock solution for LBA assays in glossaryØ Add: information regarding possibility to use “working solution” as in the chromatography

section

28

The dilution series for the preparation of the QCs should be completely independent from the dilution series for the preparation of calibration standard samples. They may be prepared from a single stock provided that its accuracy has been verified or is known.

Page 29: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

QC placement

Ø Positive clarification regarding geomeanØ Would be helpful if this was allowed also for chromatography

29

The analyte should be spiked at the LLOQ, within three times of the LLOQ (low QC), around the geometric mean of the calibration curve range (medium QC), and at least at 75% of the ULOQ (high QC) and at the ULOQ

Page 30: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Evaluation of Accuracy and Precision

Ø Rephrase to: Within-run accuracy and precision data should be reported for each run, unless there was an obvious documented error to justify the rejection of a run(s). An overall calculation of within-run accuracy and precision for each QC level should be determined. Within-run accuracy or precision criteria do not need to be met in all runs for the assessment to be successful. Between-run precision and accuracy should be calculated by combining the data from all runs.

30

Accuracy and precision should be determined by analysing at least 3 replicates per run at each QC concentration level (LLOQ, low, medium, high, ULOQ) in at least 6 runs over 2 or more days. Reported method validation data and the determination of accuracy and precision should include all results obtained, except those cases where errors are obvious and documented. Within-run accuracy and precision data should be reported for each run. If the within-run accuracy or precision criteria are not met in all runs, an overall estimate of within-run accuracy and precision for each QC level should be calculated. Between-run (intermediate) precision and accuracy should be calculated by combining the data from all runs.

Page 31: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Dilution Linearity and Hook Effect

Ø Add: Calibration standards and QCs prepared in a matrix different from the study samples should be justified and appropriate experiments should be performed

31

The same matrix as that of the study sample should be used for preparation of the QCs for dilution.

Page 32: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Dilution Linearity and Hook Effect (continued)

Ø Dilution linearity should be demonstrated by generating a dilution QC, i.e., spiking the matrix with an analyte concentration above the ULOQ, analysed undiluted (for hook effect) and diluting this sample (to at least 3 different dilution factors) with blank matrix to a concentration within the calibration range. Dilution of samples in a matrix different from the study samples should be justified and appropriate experiments should be performed. For each dilution factor tested, at least 3 runs should be performed using the number of replicates that will be used in sample analysis. A single dilution series (including 3 dilutions within the calibration range) should be assessed once. The absence or presence of response reduction (hook effect) is checked in the dilution QCs and, if observed, measures should be taken to eliminate response reduction implemented to exclude any erroneous result during the analysis of study samples.

32

Dilution linearity should be demonstrated by generating a dilution QC, i.e., spiking the matrix with an analyte concentration above the ULOQ, analysed undiluted (for hook effect) and diluting this sample (to at least 3 different dilution factors) with blank matrix to a concentration within the calibration range. For each dilution factor tested, at least 3 runs should be performed using the number of replicates that will be used in sample analysis. The absence or presence of response reduction (hook effect) is checked in the dilution QCs and, if observed, measures should be taken to eliminate response reduction during the analysis of study samples.

Page 33: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Stability

Ø Add a clarifying statement: Whole blood stability is not typically a requirement for LBA assays.

Ø A minimum of three replicates stability QCs should be prepared and analysed per concentration level/storage condition/timepoint.

33

Stability evaluations should be carried out to ensure that every step taken during sample preparation, processing and analysis as well as the storage conditions used do not affect the concentration of the analyte.

Stability of the analyte in the studied matrix is evaluated using low and high concentration stability QCs. Aliquots of the low and high stability QCs are analysed at time zero and after the applied storage conditions that are to be evaluated. A minimum of three stability QCs should be prepared and analysed per concentration level/storage condition/timepoint.

Page 34: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Stability (continued)

Ø Delete this paragraph

34

Since sample dilution may be required for many LBA assays due to a narrow calibration range, the concentrations of the study samples may be consistently higher than the ULOQ of the calibration curve. If this is the case, the concentration of the stability QCs should be adjusted, considering the applied sample dilution, to represent the actual sample concentration range.

Page 35: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Stability (continued)

Ø For both chemical and biological drugs, it is considered acceptable to extrapolate the stability at one temperature (e.g., -20°C) to lower temperatures (e.g., -70°C).

Ø Delete due to the large volume of industry data that supports extrapolation

35

For chemical drugs, it is considered acceptable to extrapolate the stability at one temperature (e.g., -20°C) to lower temperatures (e.g., -70°C).

For biological drugs, it is acceptable to apply a bracketing approach, e.g., in the case that the stability has been demonstrated at -70°C and at -20°C, then it is not necessary to investigate the stability at temperatures in between those two points at which study samples will be stored.

Page 36: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Calibration Range

Ø DeleteØ Demonstrated in validation that samples can be diluted into the fixed validated range Ø For chromatography methods no change of method is needed when adjusting

calibration range, however, for LBA it would mean a new method needs to be established (e.g. titration of reagents, change of reagents etc.)

Ø However, the fact that the assay is validated holds true regardless of technology

36

At least 2 QC sample levels should fall within the range of concentrations measured in study samples. At the intended therapeutic dose(s), if an unanticipated clustering of study samples at one end of the calibration curve is encountered after the start of sample analysis, the analysis should be stopped and either the standard calibration range narrowed (i.e., partial validation), existing QC concentrations revised, or QCs at additional concentrations added to the original curve within the observed range before continuing with study sample analysis. It is not necessary to reanalyse samples analysed before optimising the calibration curve range or QC concentrations.

Page 37: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Partial Validation (LBA)

Ø An item for trainingØ A change in detection systems and platform may be a different labelling moiety which

would constitute a new method and not partial validationØ However, stability is not likely needed to be repeated and therefore it becomes a

partial validation

37

· A change in analytical methodology (e.g., change in detection systems, platform)

Page 38: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Analytes that are also Endogenous Compounds

Ø Industry requests to rewrite this section; it is too detailed, focussed on a single type of assay. There was broad consensus that the section should be shorter and more details/emphasis should be placed on training.

Ø Change Title to “Endogenous Homologue Compounds”

Ø Change the term to mean identical or structurally similar

38

Page 39: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Parallelism

Ø Positive changeØ No longer a routine parameter as per some regional guidelinesØ Recommended change: rather than justify absence, it should be reported when performed

39

Parallelism is defined as a parallel relationship between the calibration curve and serially diluted study samples to detect any influence of dilution on analyte measurement. Although lack of parallelism is a rare occurrence for PK assays, parallelism of LBA should be evaluated on a case-by-case basis, e.g., where interference caused by a matrix component (e.g., presence of endogenous binding protein) is suspected during study sample analysis.

Page 40: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Commercial and Diagnostic Kits

Ø If an applicant uses a kit, repurposes a kit (instead of developing a new assay) or utilises a “research use only” kit to measure chemical or biological drug concentrations during the development of a novel drug, the applicant should perform a validation to ensure …..

40

If an applicant repurposes a kit (instead of developing a new assay) or utilises “research use only” kits to measure chemical or biological drug concentrations during the development of a novel drug, the applicant should assess the kit validation to ensure that it conforms to the drug development standards described in this guideline.

Page 41: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Commercial and Diagnostic Kits (continued)

Ø Specific Validation considerations for kit assays include, but are not limited to, the following:

Ø Add: Kit components should be considered as a source of critical reagents (refer to Section 4.1.2 for Critical Reagents)

41

Validation considerations for kit assays include, but are not limited to, the following:

· If the reference standard in the kit differs from that of the study samples, testing should evaluate differences in assay performance of the kit reagents. The specificity, accuracy, precision and stability of the assay should be demonstrated under actual conditions of use in the facility conducting the sample analysis. Modifications from kit processing instructions should be completely validated.

Page 42: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Glossary: LBA specific

Ø Various suggestions for additional terms and some clarifications where terminology may not suit both chromatography and LBA

42

Page 43: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Conclusions

Ø In some cases, the draft ICH M10 draft guideline allows for added flexibility and some areas have been clarified

Ø However, ICH M10 brings with it increased requirements compared to existing guidance documents– More than just harmonisation of existing guidance– Will result in increased time, resources and cost and potentially not aligned

with the mission of ICH– Preclinical requirements should be less

Ø Further clarification of some aspects of the guideline are needed to allow the correct interpretation across industry and regulatory agencies within the ICH regions

Ø Some elements may be handled through additional training as part of the ICH process

43

Page 44: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

44

Page 45: Large Molecule View: European Bioanalysis Forum Feedback · 2015–EMA Triggers for audits of GLP studies, HC stability testing, CFDA bioanalytical method validation 2018 –FDA BMV

Acknowledgements

Ø EBF communityØ Delegates and speakers at the EBF Focus WorkshopØ EBF steering committee

45