Performance Testing of Novel Dosage Forms
Terry Way BPharm MAPS
Dissolution Science ConsultantGlasside Technologies Ltd.
© 2016 – Glasside Technologies Ltd., UK All Rights Reserved
RQA Ireland Regional Forum - Athlone, May 2016Quality Considerations – Pharma and Biopharma
0
An Overview of Dissolution
The dissolution test has evolved to become a definitive tool used to characterize the performance characteristics of many types of pharmaceutical dosage forms.
As dosage forms have become more varied over the last fifty years, dissolution apparatus has required continuous improvement and modification to provide suitable conditions for performance testing.
However, probably 99% of dissolution testing is performed on traditional tablets and capsules.
An Overview of Dissolution
2
Dissolution is not just about orally ingested products such as tablets and capsules.We also test :
suspensions and powderscoated beads and granulesointments, creams, gelstransdermal patches micro-needle patchesimplants, stentsmedicated contact lenseswound care productsbone cementpowders for inhalationchewing gums, etc….
An Overview of Dissolution
Dissolution is an important tool for characterizing the biopharmaceutical properties of a pharmaceutical product at different stages throughout its life cycle.Product DevelopmentAPI characterisation, Formulation evaluation, Stability testing
BioavailabilityIn Vitro / In Vivo Correlation, Bioequivalence
Quality ControlPass / Fail product release
Scale-Up and Post-Approval ChangesRaw materials, Formulation, Process, Manufacturing site
An Overview of Dissolution
Dissolution assesses the performance of drug products
To be effective, the test should be: PredictiveComparative DiscriminatoryReproducible
An Overview of Dissolution
The dissolution apparatus must maintain reproducible hydrodynamics:Physical uniformity and alignmentTemperature controlAgitation rateVolume control (esp. evaporation)Vibration controlPrecision
The apparatus must also meet Analytical Instrument Qualification requirements: IQ, OQ & PQ
Challenges
Low Dose FormulationsOral dosage units containing concentrations of analyte at
microgram or nanogram levels.Ocular or other mucous membrane direct applications.Inhalations
Small volumesAnalytical challengesLCMS?
Quantitation Limit – the lowest amount of analyte in a sample that can be determined with acceptable precision and accuracy under the stated experimental conditions.
Challenges
Long-term extended release formulationsContinuous release of low concentrationsDays, weeks, months….Implants and depotsTransdermals, micro-needlesDrug-eluting stentsBone cement
Challenges
Rapid release formulationsImmediate releaseseconds, minutes….Sub-lingualsGel films
Sampling challenge
Challenges
Poor solubilitySink conditionsLarge volumesMultiple volumesContinuous flow
Analytical challengeLCMS?Bi-phasic?
Variety of Dissolution and Drug Release Apparatus
Compendial Dissolution Apparatus
USP <711> Apparatus 1 & 2Nominal 1000ml Vessel2l and 4l vessels also officialOperational minimum – 400 ml Tablets, capsules, suspensions,
suppositories, chewable tablets, powders
Transdermal systems (Apparatus 5&6)
Compendial Dissolution Apparatus
Modified apparatus: Stationary Basket assembly
Basket for USP FelodipineExtended Release Tablets
Compendial Dissolution Apparatus
USP <711> Apparatus 3Nominal 300ml VesselTypically 175ml to 275ml volume10cm dipExtended release tablets,
capsules, beads, chewable tablets with glass beads added.
Modifications have been made to accommodate 1000ml vessels
Apparatus 3 – novel dosage forms
Nasal sponges Oral patches
Compendial Dissolution Apparatus
USP <711> Apparatus 4 Initially designed for poorly
soluble extended release compounds
Typical flow rates from 4ml per minute up to 16ml per minute
A closed system with a small media reservoir could reduce volume to less than 30ml
Extended release tablets, beads, suppositories and implants
Special cells for powders, implants and suppositories
Flow Cell
Piston Pump
Fresh Medium
Compendial Dissolution Apparatus
16
USP <724> Apparatus 7Also known as the “Alza apparatus”
USP Apparatus 7 has evolved to handle not only transdermal systems but other extended release products.
The apparatus is derived from the Bio-Dis (Apparatus 3) and typically uses volumes of 25 – 50ml Modifications have been made to
accommodate vessels from 5ml to 300ml
2cm dipUsed for: extended release tablets,
capsules, beads, implants, transdermals, osmotic pumps, stents and other novel dosage forms
Apparatus 7 - Variations
Apparatus for Small Volumes
Larger Volumes for Transdermals
Small Volume Drug Release Apparatus – 400-DS
Temperature probe and sampling port at bottom of the dissolution cells
• Small Volume: 5-10ml
• Low Evaporative Loss
• Use of organic solvent media
• Automated Sampling and Media Replacement
400-DS Reciprocating Holder Options
Variety of sample holdersDrug-eluting stentsMicro-particulatesMedicated contact lensesPace-maker electrodes
Micro-needle patches
Compendial Dissolution Apparatus
Semi-SolidsThe requirement for testing the release rate of drugs from semi-
solid dosage forms has increased in recent years.Products tested include ointments, creams, pastes, gels and
lotions.USP has recently published General Chapter <1724> “Semisolid
Drug Products - Performance Tests”. This provides general information on testing semi-solids including descriptions of three apparatus:Vertical Diffusion CellImmersion CellInsert for Apparatus 4
Compendial Dissolution Apparatus
USP <1724> Vertical Diffusion Cell Three versions of this apparatus are described based on the classic Franz cell.
Model A Model B Model Cvolume ~7ml ~5ml ~10ml
Compendial Dissolution Apparatus
USP <1724> Immersion Cell Method Used in conventional dissolution testers with a special vessel and “mini-paddle”. Volume of medium: 50-200ml
Compendial Dissolution Apparatus
USP <1724> Flow-Through Cell Method Used in Apparatus 4 equipment
Modifications and Non-Compendial Dissolution Apparatus
Non-Compendial Dissolution Apparatus
Based on USP Apparatus 1&2Mini-Paddle ApparatusMini-Basket Apparatus100 or 200ml vesselsOperational minimum 30mlTablets, Capsules, Powders
Non-Compendial Dissolution Apparatus
Non-Compendial Dissolution Apparatus
Oral patch in 100ml300ml vesselApparatus 2 basket at 50rpmDirect UV measurement with
fibre-optic probes90% dissolved in 10 minutes
Non-Compendial Dissolution Apparatus
Modifications for 5l Vesselsfor routine testing of low-
solubility formulations
Non-Compendial Dissolution Apparatus
Vaginal rings are often tested in swirling flask apparatus in incubators typically in 250ml saline at
130rpm for 12 hoursManual sampling
Non-Compendial Dissolution Apparatus
Medicated chewing gums have become popular especially for Nicotine. Ph. Eur. describes one apparatus in 2.9.40 - DissolutionTest for Medicated
Chewing Gums“The gum is artificially chewed by the horizontal pistons, and the vertical piston ensures that the gum stays in the right place between chews.”
Volume of medium ~20ml“60 chews per minute”
USP has an Advisory Panel looking into a suitable procedure.
Non-Compendial Dissolution Apparatus
Inhalations There are currently no requirements for dissolution testing of particulates from
MDI/DPI aerosol products. Various devices have been used for research purposes:
Powder cell in Apparatus 4 using whole formulationCollected fractions from Anderson Cascade Impactor: filters from each fraction placed between membranes
in a special cell for Apparatus 4
Using NGI apparatus, special collection cups are covered with a membrane and dissolution measured in Apparatus 2:
Thank You!